Checking Kendrick laser's collimation

Started by adamsdp, 01/07/2006 08:28PM
Posted 01/07/2006 08:28PM Opening Post
I just received a Kendrick 2" laser collimator and made a make shift V-block to check it's collimation. As I spin the laser collimator 360 degrees in the V-block, the laser dot moves up and down about 3/4" at a distance of 15 feet.
I am wondering how accurate the collimation of the laser should be at this distance and if the amount of movement I measured is enough to affect the collimation of the telescopes I will use it in (14" f4.7 Tscope). Thanks!
Posted 01/07/2006 08:53PM #1
Not good. At 15 feet, the dot (or pattern) should remain in the same place. I'd even go for a farther distance, and play around with the collimation screws until it stays put.

Hugh Bartlett

"Praise the Lord for the expanding grandeur of creation, worlds known and unknown, galaxies beyond galaxies, filling us with awe and challenging our imaginations." 2007 Reform Siddur
Posted 01/07/2006 09:34PM | Edited 01/07/2006 09:35PM #2
I have an article that includes the Kendrick 2" Laser that explains the collimation.
http://www.astromart.com/articles/article.asp?article_id=96
It shows where the collimation screws are located.
It is not so intuitive, so move a screw a small amount and check the result. Only move one screw at a time while collimating. I also include a printable target for the wall that helps to track the movement and allows you to make a mark while tracing the movement.
Hope this helps.

[SIZE="Large"][/SIZE][COLOR="Blue"][/COLOR] Floyd Blue grin
Amateur Imager